Wed Nov 27, 2013 9:16 pm by mleipold
Hi Hugo,
1. Yes, all atoms are ionized with the same efficiency in the plasma. The difference in signal intensity between, say, La139 and Tm169 is due to the mass filtering of the quadrupole mass selector. The quadrupole "kicks out" masses not in the desired mass window; it does this both above and below the mass window. With its current ~AW103-203 mass window (for a CyTOFv1), it does the best job retaining the ions near the center of the mass window. This is why there's the peak of greater "brightness" near the middle (well, shifted slightly to the higher end of the middle, as Tm169 is generally the brightest). You can imagine that in throwing out AW102, it throws out a lot of AW103. Same on the high end: to throw out basically all AW204, some AW203 gets thrown out too.
The mass window is tunable (it's based on the trigger delay), so if you shifted the mass window to 120-220, the "brightest" region would also shift higher (from Tm169 to maybe Lu175). If you shifted it down, the peak would shift down.
CyTOFv2 has a broader mass window (AW78-212, according to Scott Tanner at CYTO2013), so its optimum brightness may be at a slightly different position than a regular CyTOFv1
2. The cell transmission efficiency is primarily due to the fluidics and the spray chamber. The cells stick to the sample loop tubing, the valve, the nebulizer tubing, the nebulizer, and also build up in the spray chamber, ball joint, and the injector. If everything is clean and optimized (coaxial stream/spray from the nebulizer, properly adjusted makeup gas and nebulizer gas flow rates, etc), your cell transmission efficiency should be in the 20-30% range. However, if your nebulizer is partially clogged and is therefore spraying off at an angle, this will clearly lower your cell transmission efficiency.
I would like to make a related point: if the nebulizer or nebulizer tubing are partially clogged, your tuning solution signal will usually be lower. This is also a spraying efficiency issue. However, the tuning solution is *free metal* in solution: if half your tuning solution winds up on the walls of the spray chamber, your ion count/signal will be low.
However, if you then start running cells, the *cell-associated* signal will still give you the correct signal intensity for the cell.....you'll just have fewer cells making it into the machine. This is an important distinction: you won't be losing half your ions/signal associated with a cell (and therefore have an artificially "dim" cell), just losing half your cells.
Mike