Hi Sina,
Some of that comes down to your initial panel design:
1. Do you know ahead of time (previous literature, etc) which markers are likely to be really bright? If so, did you put them on dimmer metals, and dimmer markers on brighter metals?
- note: this includes SBio surface marker conjugates. Do not assume that SBio has done a perfect job of matching marker expression with metal brightness.....you always have to check in your case.
2. Even if you have several brighter markers, you can usually find ways to mitigate spillover.
- one side comment: the "M+1" and "M+16" nomenclature that a lot of people use is *not* completely correct, and can lead to misconceptions.
- M+16 is always related to M: M+16 is the oxide spillover of M. The magnitude of M+16 spillover is directly related to the signal at M, factoring in how easily oxidized M is. The trend is that low lanthanides (La and Pr, but also Nd and Sm) are more easily oxidized than high lanthanides (Er, Yb). This is why the Tuning procedure checked the oxidation signal due to La: if you keep LaO below 3% of La, then you're automatically keeping all the other M+O oxides below 3% of M.
* however: "M+1" (related to isotopic impurity) should really be thought of as "the next highest isotope in that element", which is *not* always the same as the next increment in mass. For example: M=150 is an Nd isotope, while M=151 is an Eu isotope. Nature doesn't make a 151Nd, so 150 should never spill into 151 due to isotopic impurity. And in this case, vice versa, since nature doesn't make a 150Eu either. In the case of 151Eu, the isotopic spillover you would be looking for would be 153Eu (=M+2)
So, you can use Element (Pr, Nd, Sm, Eu, Gd, etc) choice to help with some of the spillovers. Similarly, the monoisotopic elements (Pr, Tb, Ho, Tm) can also be used to help control this: since they're 100% that channel, they won't spill into neighboring channels due to isotopic impurity. And, there's nothing else that will directly spill into them either: eg, 159Tb is the *only* naturally occurring 159 mass.
3. You may be able to use some mutually exclusive marker in adjacent channels. For example, you may have prior knowledge that a particular methylation marker is mutually exclusive of a particular acetylation marker (eg, H3K27me1 vs H3K27ac).
** To directly answer your question: panel design is iterative. I personally tend to spend a lot of time in initial panel design, choosing marker expression levels and appropriate metal brightness. I feel that gives me the best chance of making a good panel the first time around, which then allows me to titrate my full panel at once.
Once I feel I have a good titer on everything, *then* I would start checking on some spills (next isotope above and below, M+16, etc) for things that should be biologically negative. The results of that would potentially cause me to change my titer a bit to reduce spill, or potentially require me to make a new conjugate or two and retiter them.
Ultimately, a Metal Minus One (MMO) is the surest way to be check on *total* (all kinds of) spill into a channel. But yes, if you want to do some carefully chosen Metal Minus Many (MMM), then this paper from Bill O'Gorman's group has some good tips:
https://doi.org/10.1007/978-1-4939-9454-0_3Mike