Tue Jan 16, 2018 5:53 pm by mleipold
Hi Derek,
The Helios instruments are meant to be left powered on to maintain vacuum in the ion optics (quadrupole, TOF, etc) sections. It will take a minimum of a few hours to get the vacuum to the correct levels after shutting it down; overnight is often better to make a really good vacuum. Also, depending on how long it's been turned off (and exposed to atmosphere), part of that time will be getting moisture out of the system. The HIMC only turns off their instruments as needed for service calls, otherwise we leave them powered under vacuum even over holiday breaks.
The instruments are generally happier if they see a certain amount of use.....I'd say 2-3hr per week, if possible. So, if you turn on, warm up, Tune, clean, then run EQ beads and then clean again, you'll probably be pushing the 1.5-2hr mark, which is probably sufficient.
I personally find that it's worth testing the Fluidigm antibodies: most of them will be 1 uL=1 test, but some will be slightly over- or under-titered at that point. I think everything I've worked with has been in the 0.5-1.25uL per test range. I will say: we've seen some differences in titer between experiment types. For example, live-cell assays (live surface phenotyping, live cell surface staining during intracellular cytokine staining) sometimes have different titers than fixed-and-perm'd assays like phosphoflow, especially for things like CD45 and CD45RA. So, it's probably worth an experiment to test that out. For the Fluidigm conjugates, my experience is that they remain at the found titer (ie, no need to retiter between tubes or even lots). But I haven't tested that exhaustively.
If you're not using the Helios very often, then argon supply is something you'll want to consider. Liquid argon low-pressure dewars are cheaper on an hourly basis, but will exhaust and vent themselves just sitting in the corner. Conversely, compressed gas argon cylinders are more expensive on an hourly basis (one 2500psi cylinder is typically 6hr of runtime for *one* machine*), but you can leave them in the corner for months or years without losing any argon. Our typical advice to new operators is that somewhere around the 20hr (3cyl), 2-3 days/week mark on a regular basis is where you usually want to switch to liquid argon.
Mike