Hi James,
There are two issues around event rate.
1. Clogging. Too concentrated of a sample, even polystyrene beads, will increase the likelihood of clogs. For healthy donors, cryoPBMCs can be run a bit faster than whole-blood or bone marrow, which can be run a bit faster than tissue dissociates.
2. Doublets. Too concentrated of a sample, and you'll be throwing out most of your events as doublets. Remember, due to charge repulsion, the ion cloud originating from a cell expands as it travels through the plasma and into the cones. For the HT injector (original narrow-bore original Helios, which I and many others still use), I've heard numbers of around 100x (10um cell -> 1mm ion cloud).
Therefore, it's entirely possible to have cells that are separated while in suspension but whose ion clouds overlap upon charge reuplsion expansion. These events would likely then be read as *one* cell event, not two, and thrown out in various debarcoding and/or Event Length trash removal steps.
As such, one primary way in which Doublet rate is controlled is by having more dilute cell suspensions. We typically run even cryoPBMC at 0.7-0.8M/mL, for a target event rate of 250 events/sec.
Back when we first got the Helios instruments (a bit over 5 yr ago, how time flies), I did do a formal check on recovery, clogging, and doublets. Unfortunately, I can't seem to find that Excel in Cytoforum at the moment. I did discuss some of it here:
viewtopic.php?f=1&t=382&p=1336Mike