Hi Chad,
Are you thinking about spiking the beads in at the *beginning* of the protocol (like TruCount for flow)?
Over the years, there's been some discussion about this.
The biggest issue that keeps coming up is the cell recovery on the instrument: the best recoveries are usually 40-50% on a Helios or with a Supersampler, but inversely that means that 50-60% of your sample is missing (and that's not counting washing-based losses along the steps of the protocol).
Even under optimal circumstances:
1. All cell types going "missing" at the same rate to avoid cell ratio distortions (generally a decent assumption as far as I can tell)
2. Low number of bead+cell doublets which might necessitate low cell concentration and therefore long acquisition time)
3. Well-behaved non-clumpy sample (therefore uniform data acquisition)
4. Beads behaving like cells and being "lost" in washes just like cells (might be reasonably true?)
that still gives you really big error bars on absolute counting (rather than relative counting). At the very least, I think there would be a lot of preparatory experiments to characterize your instrument's recovery (ie, that your recovery variance of CyTOF vs TC20 or whatever is tightly 44-46% rather than lower 35-40% or broader 40-55%)
Looking at the BD TruCount instructions
https://www.bdbiosciences.com/en-us/pro ... ivd.340334https://www.bdbiosciences.com/content/d ... -22402.pdfit looks like there are several important aspects:
1. Commercially prepared tubes with known amounts of beads.
- at least within your own project, you could probably make a "lot" of tubes that would be fairly internally consistent for your purposes.
2. Addition of precise amount of sample.
3. Lyse-no-wash acquisition (or at the very least, I don't see any centrifugation steps or buffer washes in the TruCount protocol).
* I'm not sure how you would do this in CyTOF without having tire tracks in your acquisition. How much super-stringent Ab titration (including longer overnight staining like in Whyte et al
https://doi.org/10.1002/cpz1.589 ) could help, I'm not sure. But I don't know that it would fix your Ir intercalator streaking.
4. Not explicitly mentioned, but implied: the much higher cell recovery on a flow cytometer.
I've never used TruCount on flow, so anyone with hands-on experience, please chime in and comment on how they would see a CyTOF version working!
Mike