CyTOF XT Performance in comparison to Helios
As the CyTOF XT has been around for a couple months, I'm just starting a thread so that we can all share our experiences with the CyTOF XT, how it compares with the Helios, if you've had any troubles with the new XT and how you've troubleshooted those problems, what's easier and worse at a user/operator level than the helios, etc.
As for myself as a user, running complex epithelial cell samples and with co-cultures with fibroblasts, the XT is much more sensitive to becoming blocked (esp the nebuliser) than the Helios+Supersampler combo. The clearing clog procedure is not used at all - we've found it just causes the nebuliser to become blocked (with our samples at least). So we always stop this function and try to clear it ourselves using bleach or removing the nebuliser and manually removing the block - which works (most of the time). We've realised that we need to keep an eye on the XT much more than with the Helios+Supersampler combo, ensuring the pressure doesn't increase too much.
We've also observed that the XT calibrates/tunes the 190+ mass range differently to the helios. My reason for saying this is because we use the Cisplatin range available to us (i.e. the 194-/195-/196- & 198-Pt) as a live/dead cell stain but also for barcoding (I make the TOBis barcodes myself in-house). We also use the 191-Ir and 193-Ir channels for DNA staining. We've observed that the absolute intensities for all the stains I've mentioned have decreased by literally one order of magnitude. So the iridium intensities used to be 10^4 with the helios, but now it's at 10^3. The cisplatin intensities for our barcodes used to be between 10^3 and 10^4 but now we're only getting an intensity of 10^2 - this has caused the debarcoding efficiency to decrease slightly as well. Does anyone why this could be? I'm currently in the process of QCing and adjusting the concentrations of cisplatin in my barcodes to ensure that the dynamic range between the positive and negative signals are more distinguishable as a resolution to this problem but it would be nice to know why there's a difference in the 190+ mass range between the XT and helios.
However, not all bad news. The carousel works amazingly, and I love how (nearly) everything can be done remotely - including washing and shutting down the XT.
It would be nice to hear the experiences from other users and also if they've observed the same differences in the 190+ mass range.
Best regards,
Jahangir Sufi