EDTA Concentration during sample acquisition
I wanted to ask if users add EDTA into the sample for acquisition (if you do so at all)? If so, what concentration do you add? And what is the maximum concentration that can be used which is safe for both the (1) cells and the (2) components of the machine?
We run quite complex samples on the XT (colonic organoid epithelial cells with co-cultures of fibroblasts) and our samples (more often than not) block either the tubing or the nebuliser itself, which causes a lot of stress for both us users and the operators. But adding 2 mM into our sample during our sample acquisition has made the sample introduction much more smooth.
One thing that we tried back in the helios days when the cells were acquired in water was adding 5 mM EDTA into our sample acquisition buffer, but it was stripping off the metals from the antibody. Has anyone ever observed this whilst using a high concentration of EDTA? Also, is there anything in the CAS+ buffer which could affect the activity of EDTA?
Best regards,