Hi Luka,
I'm glad that the contamination is gone. By "acid wash", do you mean nitric, Wash (HF), or both?
Yes, samples from smokers can have certain backgrounds (among others, see
viewtopic.php?f=10&t=789&p=2319&hilit=mischmetal#p2319 ), though in that case it often includes Neodymium which I don't think you reported.
Regarding clogging of WB samples: Stanford HIMC has just completed our part of the NIH COVID-19 IMPACC study, which used EDTA WB with SmartTube PROT1 buffer fixation. Mt Sinai (the other BLD CYTOF IMPACC site) developed the staining protocol based off their pre-IMPACC COVID studies (eg,
https://doi.org/10.1002/cyto.a.24317 ).
In that Cytometry A paper, they mention:
"It is important to note that the standard SmartTube thawing protocol as per the manufacturer's instructions worked well for all healthy donor samples used in our initial validation experiments; however, when applying this protocol to blood collected from hospitalized COVID-19 patients we observed several instances in which the stabilized samples appeared to be partially clotted and exhibited high amounts of debris after thawing and lysis, which we suspect may be related to polymerized fibrin or other plasma factors related to COVID-19 disease-associated coagulopathy. If not addressed, this debris contributed to overall poor sample and staining quality and in some cases precluded analysis of samples. We found that following the red blood cell lysis washes with three additional large volume washes using ~10 ml of PBS + 0.2% BSA with centrifugation at 250 rcf and followed by filtration through a 70 micron filter depleted the majority of this debris and permitted effective analysis of blood samples that would otherwise have been discarded."
Basically, after the ThawLyse steps, there are 3 lower-speed CSB (or CyFACS above) "platelet depletion" washes. It was honestly surprising sometimes how much this can help clean up the sample. It obviously won't un-clot the sample, but the red clotted material seems to remain suspended somewhat at the lower speed (250g rather than 400g) and so it can be aspirated away. So, your cell recovery will still be low (whatever fraction of sample that didn't clot remains), but at least the clumpy material will be greatly decreased and hopefully help your samples run more smoothly.
Mike