Post Fri Mar 06, 2020 7:02 pm

Re: Iodine Contamination

We were having similar issues with Iodine contamination of samples and the EQ bead bottle. Discovered Iodine was coming from the pipette chamber, especially if pipetting up and down to mix sample. Switched to filtered pipette tips, solved the issues. Also never pipette directly from any of the stock reagent bottles.

Mike