FAQ  •  Register  •  Login

CyTOF spare parts in first 2 years of running?

<<

vtosevski

Contributor

Posts: 44

Joined: Wed Nov 20, 2013 12:50 pm

Location: Zurich, Switzerland

Post Sun Apr 13, 2014 8:52 pm

CyTOF spare parts in first 2 years of running?

Hi all,

We're just having our CyTOF being installed tomorrow in the core facility. We are now asked to estimate a budget we would need for *spare parts* withing the first two years of usage (while the instrument is still under warranty). So, my question is - how much hardware (roughly) have you used up/had on stock in the first 2 years of running CyTOF, beyond what was covered by the warranty? Extrapolation from a shorter period would help as well :)

For instance, a colleague made this list:

No. Name spare at delivery extra spares
101429    Small Air filter-Air Pump Side 1
101430    Large Air Filter- RF Gen Side 1
101519    Sample Capillary Kit 1 +2
101784    Torch O-ring Kit 1
101792    Torch Body 1 +2
101793    Injector 1 +2
101794    Nebulizer 1 +2
101803    Load Coil, CyTOF 1
101810    Vacuum Oil, HE-100, 1 Gallon 1
101814    O-ring, sampler, pack of 5 1 +1
101815    O-ring, reducer, pack of 5 1
101817    Nebulizer o-ring, pack of 5 1 +1
101829    Nebulizer port 0-rings, pk of 5 2
101801    Sampler Cone 1
101802    Skimmer-Reducer Assembly 0 +1
101425    Spray Chamber 1 +2

Can you, please, comment on this list and add your suggestions?
I understand you cannot possibly predict what I am going to break on the instrument :) but I would like to learn about your experiences on this.

Thanks a lot and best!

Vinko Tosevski
<<

mleipold

Guru

Posts: 5795

Joined: Fri Nov 01, 2013 5:30 pm

Location: Stanford HIMC, CA, USA

Post Mon Apr 14, 2014 7:57 pm

Re: CyTOF spare parts in first 2 years of running?

HI Vinko,

Here at Stanford HIMC, we have 2 CyTOFv1 machines. We're running fairly constant at about $15,000/yr for spare parts in total. So, a rough estimate might be $8,000/machine/yr.

If you add in things like spin filters and blue-capped tubes to filter your sample immediately before injection, then that rises to ~$10K/machine/yr in total.


It's really difficult to put a number on that, though: the HIMC had a machine that was having arcing problems, and so we went through Rf coils and torch bodies for a while. This FY, that's fixed, but we've gone through nebulizers faster than expected due to irreparable clogs, breakage, etc.


I have attached our FY13 cost sheet as a guide.


Mike
Attachments
CyTOF parts-non MAXPAR consumables-purchased FY13-b.xlsx
(38.33 KiB) Downloaded 473 times
<<

vtosevski

Contributor

Posts: 44

Joined: Wed Nov 20, 2013 12:50 pm

Location: Zurich, Switzerland

Post Wed Apr 16, 2014 7:29 pm

Re: CyTOF spare parts in first 2 years of running?

Hi Mike,

Thanks a lot, this is exactly what I was looking for!!

Best,
Vinko
<<

bjcarter

Participant

Posts: 2

Joined: Tue Jan 21, 2014 10:17 pm

Post Fri Apr 18, 2014 6:10 pm

Re: CyTOF spare parts in first 2 years of running?

I might suggest one other purchase that has improved CyTOF sample QC in our lab. The BioRad TC20 cell counter gives both an accurate cell count of your samples (optimizes push number/ cell recovery/ prevents overloading) and an image to access quality (cell integrity and the presence or absence of aggregates). It certainly says nothing about the quality of labeling (ie. titer) but it has limited issues with Nebulizer clogging and overloading of cells. I believe we paid around $3,000 for our unit and slides run approximately $0.50 to $0.70 per sample. -brandon

Return to CyTOF general discussion

Who is online

Users browsing this forum: No registered users and 7 guests